Karl Fischer Titration:  Coulometric vs. Volumetric

By Hank Levi on Tue, Sep 01, 2026 @ 12:34 PM

You can say we get this question all the time or we run into this situation or question all the time. 

So you may be asking yourself these questions?:

  • Why are there two types of Karl Fischer titration?
  • How much moisture can a Karl Fischer Titrator measure?
  • How much can a coulometric Karl Fischer measure?
  • How much can a volumetric Karl Fischer measure?
  • How does coulometric Karl Fischer Titration work?
  • How does volumetric Karl Fischer Titration work?
  • How do I decide which method to choose?
  • What are the benefits of using volumetric Karl Fischer Titration?
  • What are the benefits of using coulometric Karl Fischer Titration?
  • What kind or reagents do coulometric Karl Fischer Titrators use?
  • What kind of reagents do volumetric Karl Fischer Titrators use?


    # Karl Fischer Titration: Coulometric vs. Volumetric

    ## Which method is right for your sample?

This is a common question in laboratories that perform moisture content measurement: Why are there two types of Karl Fischer titration, and how do you decide which one to use?


Both **Coulometric Karl Fischer** and **Volumetric Karl Fischer Titration** measure water through the same fundamental chemical reaction. The key difference is how the iodine required for that reaction is supplied. In volumetric titration, iodine is added from a calibrated burette. In coulometric titration, iodine is generated electrochemically inside the titration cell. 

The appropriate method depends primarily on the expected moisture level, sample matrix, sample preparation requirements, and the chemical compatibility of the reagents with the sample.

## Why are there two types of Karl Fischer titration?

The two methods exist because samples can contain very different amounts of water.

A sample with trace moisture may contain only a few micrograms of water. A high-moisture sample may contain several grams of water or more. Using the same measurement approach for both situations would reduce efficiency and could affect accuracy.

Coulometric Karl Fischer is generally selected for low or trace moisture levels. Volumetric Karl Fischer is generally selected when water is present at a higher concentration. There is an overlap between the methods, so the final choice should be based on the validated operating range of the instrument and the requirements of the specific application.

## How much moisture can a Karl Fischer titrator measure?

Karl Fischer instruments can measure water from trace levels in the parts-per-million range to nearly 100%, depending on the technique, instrument configuration, sample size, sample preparation, and method validation.

As general guidance:

- **Coulometric Karl Fischer:** approximately 1 ppm to 5% water in many applications. It is commonly used for trace moisture, often around 0.001% to 1%.
- **Volumetric Karl Fischer Titration:** approximately 100 ppm to 100% water in many applications. It is commonly used for samples with moisture levels above the practical range of coulometry.

These are typical method ranges, not universal limits. A manufacturer may specify a different working range, and the usable range can change with the sample matrix, reagent capacity, sample weight, and sample preparation. 


## How does Coulometric Karl Fischer work?

In Coulometric Karl Fischer analysis, the sample is introduced into an electrolyte solution inside a sealed titration cell. The instrument generates iodine electrochemically from iodide at the generator electrode.

The generated iodine reacts with the water in the sample. The instrument measures the electrical charge required to generate enough iodine to reach the endpoint. Because the amount of charge is directly related to the amount of iodine generated, the instrument calculates the water content from current and time. [cite-15808a]

Coulometric Karl Fischer is especially useful when the sample contains a small amount of water and a very sensitive measurement is required. It can also reduce the need to handle and standardize a separate iodine titrant because the iodine is produced inside the cell. However, the cell must be kept properly sealed and conditioned. Moisture from the surrounding air, poor stirring, contaminated electrodes, or unstable drift can affect the result.

## How does Volumetric Karl Fischer work?

In Volumetric Karl Fischer Titration, the sample is added via a 10mL burette (usually)  to a solvent in the titration vessel. A Karl Fischer composite or titrant containing iodine is then dispensed from a burette until the endpoint is detected.

The instrument calculates the water content from the volume of titrant used and the titrant concentration, commonly called the titer. The titer must be established and checked according to the laboratory’s procedure and the instrument manufacturer’s recommendations. 

Volumetric Karl Fischer is well suited to samples with higher water content because the method can add a larger amount of iodine-containing titrant. It can also provide flexibility for liquids, solids, pastes, and samples that require extraction, dissolution, or other preparation steps.

## What are the benefits of Coulometric Karl Fischer?

Coulometric Karl Fischer offers several advantages for trace moisture analysis:

- **High sensitivity:** It is designed for measuring small amounts of water, including low parts-per-million levels.
- **Iodine generated on demand:** The instrument creates the iodine required for the reaction instead of dispensing a pre-made iodine titrant.
- **No routine titrant titer determination:** Because iodine is generated electrochemically, a separate volumetric titrant standardization is generally not required. The complete system should still be checked regularly with a certified water standard.
- **Small sample quantities:** Low-moisture samples can often be analyzed using relatively small sample quantities, depending on the required water mass and instrument specifications.
- **Useful for sealed-cell operation:** Keeping the titration cell closed helps limit the introduction of atmospheric moisture during analysis.

Coulometric Karl Fischer is often a strong choice for dry solvents, oils, chemicals, pharmaceuticals, and other materials where the expected moisture level is low. The sample must still be compatible with the electrolyte, and solids or pastes may require dissolution, extraction, or an oven-based sample introduction method.

## What are the benefits of Volumetric Karl Fischer?

Volumetric Karl Fischer Titration is often the better choice for higher-moisture samples and applications that require a broader sample-handling approach.

Key benefits include:

- **Broad moisture range:** Volumetry can measure samples from relatively low moisture levels through very high water concentrations.
- **Suitable for high water loads:** It is less likely to exceed the practical capacity of a coulometric electrolyte when the sample contains a substantial amount of water.
- **Flexible sample preparation:** Samples may be added directly or prepared through dilution, extraction, dissolution, or other validated procedures.
- **Useful for difficult matrices:** A suitable solvent, elevated temperature, homogenization, or an oven accessory may help address solubility and sample-release challenges.
- **Adaptable titrant strengths:** Volumetric reagents are available in different concentrations so the method can be matched to the expected water content.

Volumetric Karl Fischer may be appropriate for samples such as formulations, solvents, oils, chemicals, and other materials with moderate to high moisture content. As with any analytical method, representative sampling and complete water release are essential for reliable results.

## What kind of reagents do Coulometric Karl Fischer titrators use?

Coulometric Karl Fischer titrators use an electrolyte system that supports the electrochemical generation of iodine. Depending on the cell design, the system may use an anolyte and catholyte or a reagent designed for a diaphragm-free generator electrode.

Typical components may include an alcohol or other suitable solvent, sulfur dioxide, an organic base, and an iodide source. Specialized formulations are also available for particular sample types, including samples that may react with standard methanol-based reagents.

The correct reagent depends on the generator electrode, cell configuration, sample matrix, and application. Use only reagents recommended for the specific instrument and cell design. Commercial Karl Fischer reagent and water standard lines include formulations such as **Aquastar** and **Hydranal**, along with other application-specific products.

Common Anolytes (goes into the main titration cell)  : Coulomat A, AG, AG-H, AG-Oil, AK

Common Catholytes (goes inside a 2-component inner burette if you are using one): Coulomat CG, CGK

See our list of coulometric Karl Fischer reagents here

## What kind of reagents do Volumetric Karl Fischer titrators use?

Volumetric systems generally use:

- An iodine-containing Karl Fischer titrant
- A solvent containing sulfur dioxide and an organic base
- A one-component reagent that combines the required ingredients, or a two-component system in which the titrant and solvent are supplied separately
- Special solvents or titrants for samples that react with methanol or create other side reactions

Methanol is commonly used in Karl Fischer chemistry, but it is not suitable for every sample. Aldehydes and ketones, for example, can react with methanol-based systems and produce additional water, which may cause an elevated or incorrect result. In these cases, a suitable methanol-free or otherwise specialized reagent system may be required.


Always review the reagent safety data, compatibility information, and manufacturer instructions before use. Karl Fischer reagents and solvents may be flammable, corrosive, or harmful if inhaled or absorbed through the skin. Use appropriate laboratory controls, personal protective equipment, ventilation, storage, and waste-disposal procedures.

Volumetric Karl Fischer Titrators work with these common reagents called "Composites", "Titrants",and "Solvents".  It's worth noting that titrants work with "special solvents" designed only to work with the titrant.

Hydranal Composite 5,2, or 1 is delivered into the solvent from a burette.  The number of the composite tells you the amount of H2O that the composite can neutralize.  5 means the composite with neutralize 5mg of H20 per mL of the composite. etc, etc, (So it is a measure of strength).  

As an alternative to "Composites" you also use "Titrants". The main difference between Composites and Titrants is that the Titrants have removed some of the characteristics from their make up unlike the Composites.  In this case, the Titrants have some of their chemical components inside a special "solvent". So in summary, Composite are complete and dose into regular solvent.  Titrants are not complete chemical configurations where some of their chemical composition is located inside a special solvent.  Why are titrants this way vs. composites?  The answer is that by separating some of the chemical components in titrants extends the life of the titrant.  By not mixing all the chemical components it delays the mixing and aging of the titrant. 

Finally, when it comes to working with components and titrants you need to be aware that the strength of these will change over time so it is require and important to run what they call a "factor" check to update their strength.  Yes, you have to input the strength of the composite or titrant into the titrator when you first start using them and then periodically check and update their strength.

See our list of Volumetric Karl Fischer reagents here

## How do you decide which method to choose?

Use the following questions as a starting point:

### 1. What is the expected moisture level?

Choose coulometry when the sample is expected to contain trace or low moisture. Choose volumetry when the sample contains a moderate or high concentration of water. If the expected level falls within the overlap, compare the instrument’s validated range, sample size, and required precision.

### 2. How much water will be introduced with each sample?

The total mass of water added to the cell matters, not only the percentage of water in the sample. A large sample with a low moisture percentage may introduce more water than a small sample with a higher percentage. The electrolyte or titrant capacity must be considered.

### 3. Does the sample dissolve completely?

Incomplete dissolution can prevent all of the sample’s water from reaching the reaction. Consider a suitable solvent, extraction, homogenization, heating, or an oven method when direct analysis is not appropriate.

### 4. Could the sample cause a side reaction?

Aldehydes, ketones, reducing substances, strongly acidic or basic materials, and other reactive matrices may interfere with Karl Fischer chemistry. Select a compatible reagent system and validate the procedure for the sample.

### 5. Can the laboratory control moisture contamination?

Dry glassware, syringes, solvents, sample containers, and titration cells are important for both techniques. Conditioning the cell, monitoring drift, maintaining effective stirring, and keeping the cell sealed when required are especially important for trace-level coulometric work.

## Coulometric or volumetric: a quick comparison

| Consideration | Coulometric Karl Fischer | Volumetric Karl Fischer Titration |
|---|---|---|
| Typical application | Trace and low moisture | Moderate to high moisture |
| Typical working range | About 1 ppm to 5%, depending on the system | About 100 ppm to 100%, depending on the system |
| Iodine source | Generated electrochemically | Added from a burette |
| Primary calculation | Electrical charge, based on current and time | Titrant volume and titer |
| Titrant standardization | Generally not required for generated iodine | Required for the iodine titrant |
| Sample handling | Often syringe addition into a sealed cell | Direct addition, extraction, dissolution, or other validated preparation |
| Main concern | Atmospheric moisture, drift, cell condition, and electrolyte capacity | Titrant stability, titer, sample preparation, and endpoint performance |

## Final considerations

Neither technique is automatically better for every sample. The best method is the one that matches the expected moisture content, sample chemistry, sample preparation requirements, instrument range, and laboratory quality system.

Accurate Moisture Measurement depends on more than selecting coulometric or volumetric analysis. Proper sampling, solvent choice, reagent selection, electrode condition, stirring, cell conditioning, drift control, and routine verification all contribute to dependable results.

Scientificgear provides Karl Fischer Titrators, Karl Fischer Reagents, Karl Fischer Water Standards, and technical support for moisture analysis applications. If you need help evaluating a method, selecting compatible equipment, or improving an existing Karl Fischer procedure, contact our team for technical guidance.

                            Request a quote for a Karl Fischer Titrator

 

 

 

 

Continue Reading

Troubleshooting High-Moisture Sample issues during Volumetric Karl Fischer Titration

By Hank Levi on Tue, Dec 23, 2025 @ 01:31 PM

Recently, we worked with a customer who owns one of our Volumetric Karl Fischer Titrators (MKV-710) and was experiencing unexpected issues when testing samples with high moisture content. Specifically, the customer observed that certain samples appeared to max out at around 1% moisture, after which the titrator would go into over-titration. Interestingly, this behavior only occurred with specific sample types.

To get to the root of the issue, we stepped back and reviewed the fundamentals of volumetric Karl Fischer titration. This back-to-basics approach helped clarify what was happening—and revealed several practical troubleshooting steps that may help others facing similar challenges.


Understanding the Basics

The MKV-710 Volumetric Karl Fischer Titrator is designed to measure moisture content from trace levels all the way up to 100% water, so measuring above 1% moisture should not be a limitation of the instrument itself.

Reagent Factor Check

The customer provided data showing that the Karl Fischer reagent factor was properly calculated using Honeywell Hydranal Composite 5, with a factor of:

5.0796 mg/mL

This is a very reasonable value and indicates the reagent and factor calculation were done correctly. Great start.


What Causes Over-Titration?

Over-titration occurs when the instrument delivers more Karl Fischer reagent than is required for the amount of water actually present in the titration cell.

Here’s how the process works:

  • The detector electrode senses moisture in the titration vessel.

  • It signals the titrator to dose Karl Fischer reagent (Composite 5).

  • The reagent reacts with and neutralizes the water.

  • Once all moisture is consumed, the electrode stops detecting water.

  • The instrument then calculates total water based on how much reagent was used.

If the electrode receives inaccurate information, the titrator may continue dosing reagent—leading to over-titration.


The Math Behind the Measurement

Understanding the math helps clarify what the instrument is doing:

  • Composite 5 reagent neutralizes approximately 5 mg of water per mL

  • 1 mg = 1,000 micrograms

  • Therefore, 1 mL of reagent ≈ 5,000 micrograms of water

  • All Karl Fischer titrators measure moisture in micrograms

Key Conversions

  • 10,000 ppm = 1.0% moisture

  • 5,000 ppm = 0.5% moisture

  • 1,000 ppm = 0.1% moisture

Moisture Calculation Formula

 
Moisture (PPM or %) = Water detected (µg) / Sample size (g)

As long as the sample size is known, the moisture result can be accurately calculated from the amount of reagent consumed.

                             Learn more about Volumetric Karl Fischer Titration


Common Causes of Over-Titration

If the instrument is overdosing reagent, a few common issues may be responsible:

1. Detector Electrode Issues

A faulty or aging detector electrode may not correctly sense when all moisture has been neutralized, sending incorrect signals to the titrator.

2. Poor Sample Mixing (Very Common)

If the sample does not dissolve or homogenize properly in the titration solvent, moisture can exist in localized “pockets.” This can confuse the detector electrode and cause continued dosing.

Solution:
Improve sample mixing and solubility. In some cases, adding a co-solvent such as xylene can significantly help break down the sample.

Pro Tip: Xylene can be added at up to 20% of the total sample volume to improve solubility.


Additional Pro Tips for Reliable Results

  1. Verify Instrument Performance
    Run a water check using a micro-drop of pure water. A recovery of 99% or higher confirms the titrator can accurately measure high moisture levels. When combined with a successful factor check, this strongly indicates the instrument is working correctly.

  2. Increase Pre-Stirring Time
    If samples have solubility challenges, adjust the method to stir for 60 seconds before titration begins. This simple change can dramatically improve consistency.

  3. Use Mixed Solvents When Needed
    For stubborn samples, try a solvent blend such as 80% methanol / 20% xylene to improve dissolution.

  4. Watch for Chemical Interferences
    Samples containing ketones, aldehydes, or problematic pH levels can interfere with Karl Fischer reactions and should be evaluated carefully.

                                      Learn more about Volumetric Karl Fischer Titration


Final Thoughts

When high-moisture samples appear to “max out” or cause over-titration, the issue is often sample-related rather than instrument-related. Verifying the basics—reagent factor, instrument performance, electrode condition, and sample mixing—can quickly narrow down the root cause.

A systematic troubleshooting approach not only resolves the immediate problem but also builds confidence in the results moving forward.

Continue Reading

Introducing our technical Web Helpers!

By Hank Levi on Mon, Nov 23, 2020 @ 01:58 PM

Hello!  Please meet our new Web Helpers!  Moisture Elle, Karl Fischer, Shaker Sam, Terri Tenso, and Vivian Visco!  Each of our helpers will provide technical information ranging from general topics to more in-depth discussion pertaining to; Moisture testing, Karl Fischer Titration, Particle Size Analysis, Surface Tension, and Viscosity.

 

 

Continue Reading

How an oil evaporator works with a Karl Fischer Titrator

By Hank Levi on Mon, Aug 26, 2019 @ 11:13 AM

Evaporators configured for use with Karl Fischer moisture testers can provide added benefit when testing for moisture in samples that are not easily solubilized in the titration cell.  In some cases even if it's possible to solubilize the sample in the titration cell you may find that it is better to use an evaporator to improve repeat testing by keeping the cell clean.  While traditonal Karl Fischer moistutre testing is performed using a direct injection method where a syringe and needle is used to introduce the liquid sample into the titration cell, the use of evaporators follow a different approach. 

Within the evaporator family there are SOLIDS EVAPORATORS that can range from 0 to 300 degree C as well as high heat solids evaporators that can range from 0 to 1,000 degree C. 

A typical OIL EVAPORATOR will have a range from 0 to 300 degree C. 

Solids evaporators will typically be used to measure moisture in these types of samples:

  • plastic pellets (pte, etc)
  • plastic parts (you can cut into smaller pieces with a pair of scissors)
  • paper material (transformers)
  • thick grease samples
  • thick petroleum based samples like heavy paints, etc
  • and the list goes on

Oil evaporators will generally be used when working with heavy crude oil samples with high turnover testing (keeping the cell clean).

In this video we will discuss and show how the oil evaporator works with a Karl Fischer Titrator.

 

Talking points include:

  • Karl Fischer Titrator
  • Oil Evaporator
  • Base Oil
  • Bubbler Tube
  • Nitrogen Gas Source
  • Regulator
  • Thermocouple
Continue Reading

What happened to Hydranal Karl Fischer Reagents?

By Hank Levi on Sun, Nov 05, 2017 @ 08:12 PM

There has been some confusion in the market place when it comes to purchasing Karl Fischer reagents and water standards.  Some of you who have traditionally ordered the Hydranal brand from Sigma Aldrich (and it's channel partners and dealers) are discovering now that the Karl Fischer reagents and waters standards are slightly different - If you order them from Sigma Aldrich.  The names on the bottles are kind of the same, but not exactly.

milliporesigmalogo.png

Some background - what happened to Sigma Aldrich

As many people know Sigma Aldrich for many years handled the Hydranal line of Karl Fischer Reagents and water standards.  Recently (within the past year and a half or so), a company called EMD who controls the other Karl Fisher reagent line called Aquastar purchased Sigma Aldrich.  As a result of this transaction Sigma Aldrich is now part of another company and is referred to as MilliporeSigma.The Sigma Alrdrich you used to know no longer provides the Hydranal brand but intstead now offers the Aquastar line of Karl Fischer Reagents and water standards.

What happened to the Hydranal brand of Karl Fischer reagents?

honeywelllogo.jpgA well-known company called Honeywell purchased and now controls the Hydranal line of Karl Fischer Reagents.  You can still order Hydranal brand coulometric Karl Fischer reagents and volumetric Karl Fischer reagents via Honeywell and it's channel partners and distributors.

What does this mean to you and how might this effect you?

As you can imagine this has triggered a lot of phone calls with questions from end-users to purchasing agents with questions like:

  • Are these reagents the same?
  • Can I use these reagents interchangeably?
  • What are the differences?
  • Hydranal vs. Aquastar?

Here's what we know

As a manufacturer and distributor of coulometric Karl Fischer titrators and volumetric Karl Fischer titrators for many years, we have worked with both the Aquastar and Hydranal brands extensively.  We are not here to tell you one is better than the other.  We can tell you that different model Karl Fischer titrators seem to work differently depending on the brand of Karl Fischer reagent used.  It might simply be an operators familiarity with working with one brand vs. the other or it might be because a specific application just works better with a certain Karl Fischer reagent.  

-While the design is that either of these brands should work we find subtle differences in chemistry-  

Are there any real key chemical differences between Hydranal and Aquastar?

To be fair we have not reached out to Aquastar for their perspective yet (we will - and update this post).  From the Hydranal technical center we have obtained some information about what they see as strengths in the Hydranal brand and some differences with the Aquastar brand.  Here are a few notable points:

1. Different performance.  Yes they do perform differently.

2. Different solvents are used.  Example:  Aquastar Combititrant is a copy of the old version of Hydranal Composite 5 where there was no 2-Methylimidazole.  Without 2-Methylimidazole, you can have reagent crystalization and stability issues.

3. Co-solvents are optimized in Hydranal.  Too much Chloroform can mess up the stoichiometry

4. Hydranal coulometric reagents are more unique

  • Hydranal has more buffering capacity
  • Larger volume of samples than Aquastar

5. Hydranal is very precise, convenient and consistent

As mentioned above we hope to get some additional information on the Aquastar line so we can share with you their strengths and differences.  In the meantime we hope this information has been helpful.

If you have any questions or would like to compare the reagents you currently use with the "other" brand just let us know.  We can provide a cross reference for all of the Karl Fischer reagents by brand and product.

Learn More

As always, we hope this information is helpful.

 


 



 

 

 

 

 

 

 

Continue Reading

Review of KEM's new titration line

By Hank Levi on Tue, Jun 09, 2015 @ 04:11 PM


Review of KEM's new titration line 

AT-710S_M Automatic Titrator

I don't know about you but I have to admit that I enjoy learning about new technology.  Whether it's smart phones, cars, or kitchen appliances I think it's fun to find out about how a product has changed (hopefully for the better). For the most part a lot of consumer products see a change or upgrade about once every year. This is not really the case for titrators where maybe it's only every 6 to 8 years before we see a new generation unveiled.  I guess that's why I thought maybe it would be a good idea to take a look at this new 710 series that came out in February 2015.  I haven't actually ever written a "review" before so I decided to start by comparing what I know about some of the current titrator models and identify the things that pop out with the new titration line.   Here are my observations and comments.


  

1. There are now three model levels; entry, mid-level, and flagship within each category (volumetric Karl Fischer, coulometric Karl Fischer, and automatic potentiometric titrator) 

Find more Coulometric Karl Fischer Titrator information

My comment: I like the scale of the models as it can be helpful for those with varied budgets and needs.  I like the concept that you can start with getting only the basic model at first and then later bring in the flagship model and tie them all together into one system.  It will be interesting to see how people approach these options and whether the ability to expand a system is desirable or not.
710_series_titrator_with_2_burettes

2. All of the automatic titrators now can accommodate two burette drives

 AT-710 AutomaticTitrator information

My comment: The automatic titrators come with one burette drive but being able to add a second burette drive allows an operator to run two separate titrations (titrants) without having to use an automatic piston burette.  Historically most titrator manufacturers offered only a single burette drive but it seems like this is changing.  It's kind of a big deal for those who run two different titrations and don't want to purchase an automatic piston burette.  I like this new capability.

3.  A new burette design for the automatic titrators and volumetric Karl Fischer titrators.

SmartBuretteUnit

My comment: The burettes appear they will be easier to store and swap as the titrant bottle, burette and nozzles all go together as one smaller unit.   The burettes are now "smart" burettes and can store all of the reagent information within the head of the burette.  I like this.

Learn more about Volumetric Karl Fischer Titration

4.  More Input/Output options for operators to move  and store data.  I/O options include LAN connectivity/URL, USB hubs, USB thumb drive ports, .CSV file format, .PDF file format, barcode readers, foot switches, and keyboards.710seriesUSB

My comment:  More technology is built into these units. I think it was overdue so I am glad to see it finally arrive.

AT710_w_propeller_or_magnetic_stirrer

5.  Automatic Titrators now come with a propeller stirrer by default but can swap for a magnetic stirrer instead.

My comment: I like the option to do both but I think maybe the magnetic stirrer should have been the default and the propeller as the option.

 

6.  A new Wireless/ Wired 8.4 inch color touch panel controller (MCU).

MCU_screen_view_with_multiple_units

My comment: This only comes with the mid-range and the flagship models.  The controller can sit next to the titrator or be carried around like a clipboard.  Operators can control the titrator through protective glass if necessary.  I'm not sure how many people will elect to use the wireless vs. the wired connection but it's nice to have the added capability.  Nice.

7.  Multi-titrator integration.  The wireless/wired MCU can control up to 4 different titrators simultaneously (any combination of Karl Fischer or automatic titrator) 

MCU_controller for 710 series titrators

My comment: This feature is reserved for the flagship model.  I can see this being useful for those needing to run both moisture and acid for example.  Unfortunately the mid-level model MCU cannot be upgraded to the MCU flagship model so if you think you might expand later don't go with the S model.  Your better off getting the basic "B" model and then later tying them all together from the flagship model.

 710seriesmulticontrol

8.  Free method making software.

My comment:  All of the titrators now come with a CD that allows you to create methods on your PC and then transfer methods to the titrator.  You can also pull methods off of the titrator and edit it in the PC.  The software also comes with various popular methods pre-loaded to help get methods setup quicker.  I'm not so sure how robust this software is but free is nice.

CDstoredmethods
 
 
 
 




CONCLUSION:

Well I hope this was helpful in finding out quickly what the updates are for the new 710 series.  At this point since it's so new we will have to wait to evaluate how customers feel about their user experience and get their feedback on what they liked most and what they liked least.  If you have any questions or comments you can email me at info@scientificgear.com



 

Continue Reading

How to use Karl Fischer Grease

By Hank Levi on Sun, Jan 19, 2014 @ 01:57 PM

Karl Fischer GreaseCertain types of Karl Fischer vessels require the use of Karl Fischer grease. Vessels with smooth port openings need a thin layer of the grease applied to plugs, electrodes, dessicant tubes, bubbler tubes and injection port plugs to help form a snug fit.  Decreasing or limiting "ambient moisture" from getting into the vessel - otherwise known as "drift" - is the key benefit of using Karl Fischer grease.  Another benefit of Karl Fischer grease is that it also helps prevent chipping of glass on glass fittings.  Watch this short video to see how Karl Fischer Grease should be applied.

Learn more about Karl Fischer Titration
If you find this information helpful please consider subscribing to our blog.  Once you subscribe you will get notified when we post a new article.

Continue Reading

How do Ketones & Aldehydes affect Karl Fischer moisture testing?

By Hank Levi on Mon, Sep 23, 2013 @ 04:32 PM

Measuring moisture using a karl Fischer Titrator is a common and popular choice among operators.  Karl Fischer titration provides excellent accuracy at the part-per-million (PPM) level and is widely accepted and trusted as a reliable method for measuring moisture.  Ketones and Aldehydes however are two organic compounds that can cause problems and lead to errors with measured results if they are present in a sample during testing.  The presence of these organic compounds can create an unwanted side-effect inside the titration vessel.

 ketoneWhat are the side-effects?   aldehyde

The effects are side-reactions that can either falsley increase your moisture results or falsley reduce your moisture results.  Depending on whether your dealing with ketones or aldehydes or both you will want to know how to deal with them.  

How do you solve the problem?

Once you have identified the problem you need to take stock in your equipment and the chemical reagents you are using.  The equipment setup and reagent choices you make can help you to suppress these side reactions and obtain accurate results.

Watch this short 3 minute video to learn about the preferred equipment setup and your chemical reagent choices.

 

 On a sidenote who knew that Ketones have an impact on the human body? Check out this cool review

Just thought it was interesting and wanted to share.

                           Still need help? Click here.

 

Continue Reading

Karl Fischer Over Titration and 7 things you should know about it

By Hank Levi on Wed, May 02, 2012 @ 06:30 PM

We have all seen it.  We’re running a test to see how much moisture is in our sample when inexplicably the liquid inside the Karl Fischer vessel starts to turn from a normal light-yellow color to a dark burnt-red looking color.  Sometimes the titrator screen will inform us of the problem with a digital readout stating the dreaded

OVER-TITRATION!” 

kfbeforestart resized 600describe the imagedescribe the imagedescribe the image

 

Sometimes the digital readout says nothing at all.  In either case the operator knows something has gone wrong because the Karl Fischer Titrator is no longer giving moisture results.  A panic to figure out the problem and get testing underway again becomes the immediate priority. 

But where do you start?  

As it turns out “over-titration” is probably one of the top 2 or 3 complaints or issues we hear about from operators.    So what is causing this problem to occur?  How can we determine the source of this problem, fix it, and more importantly how can we avoid it?

 
This is a critical question for operators and managers working in a production or QC environment who are concerned with keeping their Karl Fischer Titrator 100% “in-service”, day-in and day-out.   Having the ability to identify the problem correctly so appropriate measures can be taken quickly is very important.

 
So how do we approach the problem of “over-titration”?   By knowing the facts.  Having a clear understanding of the process can help operators correct the problem faster when time is of the essence.  

Important facts you should know about over-titration:  
  1. Over-titration is a state where there is more iodine present in the vessel than water (general definition).
  2. When over-titration occurs the vessel will become very dark as a result of the abundance of iodine present inside the vessel.
  3. The reagent inside the vessel should normally have a light-yellow color absent a very dark sample such as oil.
  4. The Karl Fischer Titrator always attempts to maintain an equilibrium where only enough iodine is introduced to counter and neutralize the water present inside the vessel.
  5. During a single titration test there should only be enough iodine introduced to counter and neutralize the amount of water present inside the titration vessel during that test – no more, no less.
  6. Any incident that interferes with the final amount of iodine introduced during the titration test can lead to dis-equilibrium and result in more iodine being introduced than necessary.
    Important facts you should know about Karl Fischer Titrator Glassware:  
  1. The Karl Fischer Vessel and Glassware is composed of the following
    1. Vessel (coulometric and volumetric)
    2. Generator Electrode (coulometric only) - The Generator Electrode is a precision electrode designed to deliver an electrical current inside the vessel to the reagent – causing the reagent to produce iodine
    3. Titration nozzle (volumetric only) - The titration nozzle delivers precise amounts of iodine (composite or titrant) via a burette driven mechanism using a piston
    4. Detector Electrode (coulometric and volumetric) - The detector electrode has a sole purpose and probably the most important role in continuously monitoring and determining the conductivity levels within the titration vessel.
  So what are the causes that can lead to over-titration? 

✓ A damaged Detector Electrode

✓ A "Tricked" or "Fooled" Detector Electrode (no joke)

Since coulometric and volumetric Karl Fischer Titrators handle the delivery of iodine differently it’s worth describing the two methods separately.  

A Word About Coulometric Karl Fischer Titration:

In a coulometric system the reagent is a complete system where it is designed to release iodine when the generator electrode delivers an electrical current to it.  So what causes the generator electrode to deliver too much current causing the over production of iodine?  Another way to say it is, “who or what” is telling the generator electrode to continue to generate a current when it’s not needed?  

   
The detector electrode!  So why would the detector electrode do this?   describe the image
Without getting into too much of the electronics the detector electrode is designed to “detect” conductivity in the vessel.  Depending on the amount of conductivity detected the detector electrode will send a message to the titrator telling it to continue producing a current - enough to release the appropriate amount of iodine to counter and neutralize the water present in the vessel.  As long as this process is working during a titration an eventual endpoint will be found and a result will be produced.  


So it’s really a problem of misinformation. If the Karl Fischer Titrator is not getting the right information from the detector electrode then over-titration is possible. 

The problems we see that can effect the proper functioning of the detector electrode include:  
  1. The electrode cable.  If the cable becomes cracked or breaks it can cause a situation where the message to the titrator is to continue producing a current – continually.  In this case the vessel will become very dark and in most cases the titrator will not even know it is in an over-titration state.  The generator electrode will simply continue to produce a current, turning the vessel very dark.  There will be no other warning or notice from the titrator for the operator to see.
  2. Cracked electrode.  Sometimes mishandling or even a stirrer bar bouncing around inside the vessel can cause a tiny crack near the bottom of the detector electrode that cannot be seen with the naked eye.  These cracks can allow small amounts of reagent inside the electrode enough where errors in detection will begin to occur.  What ensues is an unstable drift that jumps around giving the titrator a misreading. The jumping around and unstable drift may be picked up by the titrator and an error stating “OVER TITRATION” may be seen on the screen of the titrator.
  3. Cable connectors.  Sometimes the connectors on the titrator itself can become dirty, wet and corroded.  Also, some electrodes use multi-plug designs that can also become dirty, wet and corroded.   These connectors if not clean and dry can lead to a similar misreading similar to a cracked electrode where the drift begins to jump around and become unstable.  The titrator may also state that there is “OVER TITRATION” when this occurs.

Find more Coulometric Karl Fischer Titrator information 

A "Tricked" or "Fooled" Detector Electrode you say?describe the image

If it’s determined that the problem is not the detector electrode then we need to look at the stirring action inside the vessel.  If the iodine being released is not mixing well because the stirrer is off or set too low, then the detector electrode will not realize there is iodine already released inside the vessel.  This will cause the detector electrode to continue telling the titrator to produce more current via the generator electrode up to the point where the detector electrode senses a reduction in the conductivity level inside the vessel.  Conductivity only reduces as the iodine interacts with the water.  So it is important for the detector electrode to sense the true and most accurate “mix or state” of iodine and water during the titration process.  If it does not know the true state of the mix it will be fooled into telling the titrator to keep going – causing OVER TITRATION.        

A Word About Volumetric Karl Fischer Titration:

   
describe the imageIn a volumetric system the reagent setup is different where a composite or titrant is introduced via a burette piston through a titration nozzle.  The amount of composite or titrant delivered is based upon the commands of the titrator.  The command from the titrator to the burette and piston that push out the “iodine” through the titration nozzle is, yes, given by the detector electrode.  For the purposes of this discussion the difference between the coulometric and volumetric setup is that the delivery of iodine is different.   But the same problem can occur where the iodine does not mix well and therefore trick the detector electrode in to thinking there is not enough iodine present inside the vessel to counter and neutralize the water.  Since both coulometric and volumetric Karl Fischer Titrators use detector electrodes the problems mentioned earlier about the detector electrode will hold true with volumetric titrators also.  

Learn more about Volumetric Karl Fischer Titration 

 

 7 Thoughts (DOs and DON'Ts) on Problem Solving and Prevention:  

1.  Don’t abuse the detector electrode!  Be very careful with the detector electrode and do not handle it unnecessarily.  Small bumps (clanks) here and there can lead to a crack.   Do you really need to remove the detector thinking dragonelectrode from the vessel all the time?
2.  Don’t turn up the titrator’s stirrer speed to high.  This will only cause the stirrer bar to bounce around uncontrollably and possibly hit and damage the detector electrode (crack).
3.  Do inspect all connections and connectors on the detector electrode cable and Karl Fischer Titrator to ensure they are dry and clean.
4.  Do be careful with the detector electrode cable.  Try not to bend it unnecessarily.
5.  Do make sure there is enough stirring action inside the vessel to mix the iodine around effectively.  A small vortex should be visible.  But not too fast to cause the stir bar to bounce around.
6.  Do introduce some moisture - Sometimes when you are in an over-titration situation and the vessel is already very dark you can introduce a little moisture to bring the vessel back to equilibrium.  This sometimes works and immediately the vessel turns from a dark burnt-red color to a light-yellow.
7.  Do have a spare detector electrode on hand.  This little electrode seems to get over looked but plays a huge role inside the Karl Fischer Titrator vessel.

Created on 05/02/12 at 18:10:44

Continue Reading

What are Karl Fischer Water Standards and what do the numbers mean?

By Hank Levi on Thu, Feb 09, 2012 @ 01:44 PM

Most everyone working with Karl Fischer Titration at some point ends up wanting to check their instrument for accuracy and overall operational readiness.  Karl Fischer Water Standards were made to assist operators with making these operational checks.  Sometimes however we find there is some confusion about the choice of water standards available and what the numbers mean.

Learn more about Karl Fischer Titration

Karl Fischer Titration Water Standard Video from Scientificgear on Vimeo.

Hydranal water standards provide a few popular choices for both coulometric and volumetric Karl Fischer Titrators:

  • Hydranal 0.1 (100PPM ±10% error acceptance)
  • Hydranal 1.0 (1,000PPM ±3% error acceptance)
  • Hydranal 10.0 (10,000PPM)

So what do the numbers mean?  0.1, 1.0, 10.0?

Simply put, these numbers tell us the amount of moisture (H2O) that is present in 1 gram of the water standard.  The amount of moisture (H2O) is expressed as milligrams on the packaging.  Using the Hydranal 1.0 for example we say there is 1.0 milligram of moisture (H2O) in 1 gram of the water standard.  Did you know there are 1,000 micrograms per 1.0 milligram?  Yes there is.  So instead of thinking in terms of 1 milligram per 1 gram of water standard, think in terms of 1,000 micrograms per 1 gram of water standard.  Why?

Karl Fischer Titrators count moisture in micrograms!

Since Karl Fischer Titrators count moisture in micrograms it's easier to think about the water standards in terms of micrograms.  Why?

Find more Coulometric Karl Fischer Titrator information

We evaluate our water standard test in PPM

For the Hydyanal 1.0 we are looking for results within ±3% of 1,000PPM (970PPM to 1030PPM).  For the Hydranal 0.1 we are looking for results within ±10% of 100PPM (90PPM to 110PPM).

 Learn more about Volumetric Karl Fischer Titration

Don't forget this formula!:

PPM = WATER DETECTED IN MICROGRAMS/SAMPLE SIZE IN GRAMS

(For those who don't know PPM stands for Parts Per Million)

We hope this information has been helpful.